Hi Raja,

The configuration of the receiver plays a big part in the sweep speed. The biggest factors that affect sweep speed on the SA44 is span, rbw, and image/spur rejection. Using Spike, a span of 200k-500k, 1k RBW, and image rejection off, I was able to achieve a sweep speed of around 22ms. I suspect this is about as fast as the API will give you sweeps.

It would be necessary to know your measurement requirements before I could determine whether the SA44 could meet them.

You could increase measurement speed by utilizing the IQ data stream, but this would then require more signal processing (FFTs, etc.).

Regards,