  • #25351
    Roger Peterson
    Participant

    According to specifications, the BB66C should have an input impedance of 50 ohms.
    Using a 50 ohm output signal generator (Agilent 33220A), an output frequency of 10KHz is selected with an output level of -20dBm.
    The signal generator is TEED with an oscilloscope (high input Z) and the BB66C.
    If the BB66C has an input Z of 50 ohms, the scope should read a -20dB level, but it does not.
    Using just an ohmmeter, the input R of the BB66C is >1M ohm with or without power.
    Using the same above test procedures with the USB-SA44B, all is normal (as expected).
    I just want to verify that the BB66C does or does not have an input Z of 50 ohms?

    #25376
    AndrewAndrew
    Moderator

    Hello Roger,

    Thank you for calling in. Hopefully we were able to fully answer your questions. If you need anything further please let us know.

    Regards,
    A.J.

    #49166
    BruceBruce
    Moderator

    For anyone else that is curious, The BB60C has a 50Ω input impedance and AC input coupling so it won’t look like 50Ω using a multimeter but at RF frequencies it is 50Ω. The SA44B has a DC coupled input so it will look like 50Ω to a multimeter.