    Hello,

    which calibration method do you use for BB60c?

    best regards.

    #239745
    Justin CrooksJustin Crooks
    Moderator

    We have a software suite for instrument calibration. See https://solutionspagedemo.kinsta.cloud/software/signal-hound-field-calibration-software/ for more information. A typical calibration lab will have most or all of the equipment required.

    #239749
    lsullivan
    Participant

    thank you for your answer, But I am referring to self-calibration part that has been written in the API manual.
    regards.

    #239767
    AndrewAndrew
    Moderator

    The self-cal function in the BB60 API is for the BB60A only and has no effect on the BB60C.