The brand-new PN400 Phase Noise Test System is available for purchase! →

RBW and VBW

Forums › SA Series Discussions › RBW and VBW

Viewing 4 posts - 1 through 4 (of 4 total)
  • Author
    Posts
  • #228640
    flo
    Participant

    Hi,

    if i measure the signal with an open input at the SA44B with RBW at 30Hz and change the VBW to 10 or 3Hz the the signal or noise level measured with a marker increases about 4 to 5dBm compared to a VBW of 30Hz.

    I always thought, that the noise is reduced with a lower VBW?

    Can someone explain me this?

    Thanks
    Florian

    RBW Shape: Flat Top
    20kHz to 40kHz
    Video units: Power
    Detector: Average
    RefLevel: -50dBm

    • This topic was modified 7 years, 1 month ago by flo.
    #228746
    Justin CrooksJustin Crooks
    Moderator

    This has to do with the SA-series software-based image/spur rejection algorithm. Behind the scenes, it is injecting LO high side, making a sweep, then injecting low side, making a sweep, and picking the lower measurement. For this reason, the SA44B is not recommended for broadband measurements (such as broadband noise).

    If you need to make a noise measurement with the image/spur rejection enabled, set your VBW to 1/10 the RBW. You will need to compensate for energy at the image frequencies if your noise bandwidth is >42 MHz. Subtract 3 dB from the measurement for these situations.

    The alternative is a BB60C or an SM200A, which have excellent hardware-based image rejection and much lower spurious. These will provide good noise measurements without any tricks.

    #228762
    flo
    Participant

    Hi Justin,

    thanks for your answer.

    So turning off the image/spur rejection will not solve the problem?

    And if I set the VBW to 1/10 of the RBW and measure below 42MHz (I measure from 20 to 40kHz or 60 to 100kHz) I dont have to substract 3dB from the measurement.

    I will try it with the SA44B at first.

    #228774
    Justin CrooksJustin Crooks
    Moderator

    Flo,
    The image response will be 21.4 MHz above the 20-100 kHz signal. If your configuration is measuring flicker noise or similar, and you have a low frequency amplifier with, say, less than 10 MHz of bandwidth, between your DUT and the SA44B, you would not subtract 3 dB, as the noise at 21.4 MHz would be negligible compared to the noise at 20-100 kHz. You could turn off spur rejection, deal with the spurs, and not worry about VBW settings. Or you could set VBW to 1/10 RBW with spur reject on, and still get a valid reading.

Viewing 4 posts - 1 through 4 (of 4 total)
  • You must be logged in to reply to this topic.