The brand-new PN400 Phase Noise Test System is available for purchase! →

Rx Converter measuring

Forums › TG Series Discussions › Rx Converter measuring

  • This topic has 3 replies, 3 voices, and was last updated 7 years ago by Ermy.
Viewing 4 posts - 1 through 4 (of 4 total)
  • Author
    Posts
  • #230261
    Ermy
    Participant

    I have to test a 144/28 mhz rx-converter with SA44B and TG44A
    but I don’t find in Spike an Offset frequency function for the tracking generator. How you do ?
    Thanks
    Ermy

    #230290
    AndrewAndrew
    Moderator

    Hi Ermy,

    We do not have a frequency offset setting for the TG in our software. I apologize for the inconvenience.

    Regards,
    Andrew

    #230366
    Justin CrooksJustin Crooks
    Moderator

    If an approximate conversion loss (+/- 3 dB) is acceptable, you can sweep the output range with the SA44B, and then step across the input range with the TG manual controls, using MAX HOLD on a trace.

    For better accuracy, an RF power meter could be used to measure the actual TG output power at each frequency and you could subtract this from the SA44B’s reading.

    #230416
    Ermy
    Participant

    Andrew thanks for the reply and also Justin’s suggestion.
    I found, however, that the function is available in the 2.18B version of the legacy software.
    I’d like to see it in an upcoming release also in Spike.
    Thank you all

Viewing 4 posts - 1 through 4 (of 4 total)
  • You must be logged in to reply to this topic.