  • #99283
    zhubajei
    Participant

    The peak value indicated in SPIKE was lower than the input Value, so SelfTest was carried out.
    As a result, Fail was indicated by 2 points. (Attachment picture referring)
    Is repair or proofreading necessary to go back up normally?

    It is another question.
    Two peak waveforms are displayed on the both sides of input signal.
    It is possible to remove this waveform?

    Attachments:
    You must be logged in to view attached files.
    #99285
    zhubajei
    Participant

    It is self-less.
    Input signal 611.125MHz -30dBm
    Anritsu MG3642A

    #99348
    AndrewAndrew
    Moderator

    Hello zhubajel,

    I will be sending you a support email using the one provided in your profile. If this email is not correct, you can email me at aj at signalhound dot com.

    Regards,
    A.J.