The brand-new PN400 Phase Noise Test System is available for purchase! →

Spur reject produces artifact in center of spectrum

Forums › SA Series Discussions › Spur reject produces artifact in center of spectrum

Viewing 2 posts - 1 through 2 (of 2 total)
  • Author
    Posts
  • #240943
    JHoy
    Participant

    I am performing spectrum measurements in sweep mode on a very low power CW tone near the noise floor. I have come to notice that turning on the Spur Reject causes an artifact to appear at the center of the span regardless of where I am tuned. This occurs with low RBW and VBW settings at or below 100Hz. If I turn Spur Reject off, the artifacts disappear, but the noise floor rises by about 5dB. Is this normal?

    • This topic was modified 4 years, 5 months ago by JHoy.
    • This topic was modified 4 years, 5 months ago by JHoy.
    #240959
    AndrewAndrew
    Moderator

    This thread was resolved via email.

    Outcome: Disabling spur reject will raise the noise floor by ~3dB. The artifact in question is likely an internally generated spur that is within the specification of the receiver.

    Andrew

Viewing 2 posts - 1 through 2 (of 2 total)