  • in reply to: BB60C #255806
    AndrewAndrew
    Moderator

    It seems like you are talking about the noise floor of the measurement? If yes, then you can change the noise floor via 2 main ways. You can adjust the RBW. If you change the RBW to 10kHz, that will lower your noise floor by 20dB to 0dBuV. You can also change the sensitivity of the BB60C. If you lower the max input by another 5 dB to 55dBuV it will lower the noise floor by another 5 dBm. Unfortunately at 55dBuV you are at the maximum sensitivity of the BB60C, and won’t be able to lower it more via the max input control. You will need to rely on RBW again.

    in reply to: BB60C #255786
    AndrewAndrew
    Moderator

    Can you provide additional clarification? I’m not sure what you mean by “mt db in y axis automatically goes to 20dBuV”.

    in reply to: BB60C #255736
    AndrewAndrew
    Moderator

    Hello,

    Unfortunately it is not possible to set gain/atten directly in the EMC precompliance mode as you can in the sweep mode, but only indirectly through the “max input” control, which has similar behavior to reference level in sweep mode.

    I apologize for the inconvenience.

    in reply to: BB60C #255735
    AndrewAndrew
    Moderator

    Hello,

    Unfortunately it is not possible to set gain/atten directly in the EMC precompliance mode as you can in the sweep mode, but only indirectly through the “max input” control, which has similar behavior to reference level in sweep mode.

    I apologize for the inconvenience.

    in reply to: Feature Request – Trigger on Pulse Width #255715
    AndrewAndrew
    Moderator

    Thank you for the feature request! I have added it to our customer request log. I will update this thread when we release a version with this feature.

    Regards

    in reply to: USB cable option? #255667
    AndrewAndrew
    Moderator

    Michael,

    You also emailed us this same question. We have responded via email.

    in reply to: BB60C-API for ARM #255553
    AndrewAndrew
    Moderator

    Hi Scott,

    I’m sorry I didn’t realize you were also the same person we have been emailing.

    I just sent a follow up email mentioning that based on my most recent understanding I don’t think there is an issue with the processing. I think the reason you see the “peak” or “corrupted data” at the 5MS/s rate is that you are centered at 100kHz, and the 0Hz (DC) feedthrough is coming into your spectrum. The filtering in the API is not sharp enough at that sample rate to filter out the 0Hz response. I recommend a post processing filter if you want to remove the 0Hz response, or reducing the sample rate to 2.5MS/s or below such that additional filtering helps remove the 0Hz feedthrough.

    Does this sound like an accurate take on the information you have provided? If yes, then I believe the API is likely working nominally on your system.

    I look forward to your response.

    • This reply was modified 3 months, 2 weeks ago by AndrewAndrew.
    in reply to: BB60C-API for ARM #255550
    AndrewAndrew
    Moderator

    sreeder,

    Have you tested them in your configurations and verified they would not work? I have customers reporting success on the RPi5 under Ubuntu when using the Nvidia aarch64 builds. Both PiOS and Debian are similar enough I would expect them to work.

    • This reply was modified 3 months, 2 weeks ago by AndrewAndrew.
    AndrewAndrew
    Moderator

    Hi Miran,

    I can reproduce. It looks like this issue was introduced in the most recent version. I already have a fix and it should go out in a release next week. I apologize for the inconvenience.

    Here’s a link to the previous version if you want a version that should not contain this bug. You will need to uninstall the newer version manually before installing.

    https://solutionspagedemo.kinsta.cloud/sigdownloads/Spike/Spike(x64)_4_0_12.zip

    Thanks as always.

    • This reply was modified 4 months, 1 week ago by AndrewAndrew.
    in reply to: BB60C-API for ARM #255367
    AndrewAndrew
    Moderator

    Scubajosh, the RBW/VBW is limited to 1kHz minimum in this build. Additional processing is needed to reach 10Hz.

    • This reply was modified 4 months, 2 weeks ago by AndrewAndrew.
    in reply to: BB60C-API for ARM #255354
    AndrewAndrew
    Moderator

    There is now a BB60 Mac ARM API build in the latest SDK release here,

    https://solutionspagedemo.kinsta.cloud/software/signal-hound-software-development-kit-sdk/

    Please review the README in the macos_arm folder to understand it’s limitations (same as the Ubuntu aarch64 build).

    • This reply was modified 4 months, 2 weeks ago by AndrewAndrew.
    in reply to: SM435B Marker Peak Right/Left #255352
    AndrewAndrew
    Moderator

    Zitro,

    The device can only be interfaced by one application at a time. Attempting to connect to the device from 2 applications/users simultaneously will cause undefined behavior. I apologize for the inconvenience.

    in reply to: Low Spur mode–VSG60 API? #255337
    AndrewAndrew
    Moderator

    Hi Michael,

    The function vsgSetDigitalTuning allows you to control low spur mode. It has a different name, I apologize for the confusion.

    https://solutionspagedemo.kinsta.cloud/sigdownloads/SDK/online_docs/vsg_api/vsg__api_8h.html#a9dd9c01340e63f3101941122e320bbcc

    Let us know if you have follow up questions.

    • This reply was modified 4 months, 3 weeks ago by AndrewAndrew.
    • This reply was modified 4 months, 3 weeks ago by AndrewAndrew.
    • This reply was modified 4 months, 3 weeks ago by AndrewAndrew.
    in reply to: SM435B Marker Peak Right/Left #255297
    AndrewAndrew
    Moderator

    Do you have a picture of the spectrum so I can provide some specific suggestions?

    Also, if the marker does not stay at the peak, then it sounds like the input frequency is drifting, is that a correct assumption?

    If you set peak threshold to a very low value, it is essentially “turned off”.

    Did you try the peak table?

    in reply to: BB60C-API for ARM #255281
    AndrewAndrew
    Moderator

    Hi scubajosh,

    I apologize for the confusion, yes we use the aarch64 folder for the builds on the Nvidia Jetson. For APIs that have ARM Mac builds we use a folder called “macos_arm”. We have them for the SP and SM APIs, but not the BB API currently. I have confirmed we do have an open ticket to get this for the BB60 API. I don’t have a timeframe on this work right now, but I will be sure to update this thread when complete.

    in reply to: SM435B Marker Peak Right/Left #255280
    AndrewAndrew
    Moderator

    Hi Zitro,

    If you wanted to provide a picture of the screen I might be able to provide you the best settings to accomplish this. I can try to describe what I think would be the best approach based on your description.

    The Pk threshold should be lower than any of the peaks you anticipate measuring. If it is higher than your second peak, then it won’t work. If your second peaks is very close to the noise floor (doesn’t rise out of the noise floor much), you could reduce the Pk Excursion from 6 to something smaller.

    With that setup, I would then press “Peak search” and then “Next Peak” to find the second highest peak.

    Another alternative is to use the peak table to show all peaks simultaneously. You can find the peak table in the control panel along the bottom of the screen while in sweep mode.

    If none of the suggestions above work, please provide a picture of your spectrum.

    in reply to: BB60C-API for ARM #255257
    AndrewAndrew
    Moderator

    There are currently no plans to port Spike to ARM at this time.

    Thanks

    in reply to: Mean-Time Before Failure MTBF of BB60C/D #255255
    AndrewAndrew
    Moderator

    Tuong,

    Can you please send this request to support@signalhound.com.

    Thanks,

    • This reply was modified 4 months, 4 weeks ago by AndrewAndrew.
    in reply to: Direct ADC (BB_DIRECT_RF) Mode ? #255227
    AndrewAndrew
    Moderator

    I apologize for the issues and confusion. At this point BB_DIRECT_RF is deprecated. We now have the ability to stream I/Q data anywhere in the frequency range of the BB60 and exclusively recommend doing that.

    I will update the documentation to reflect this.

    Let us know if you have follow up questions.

    in reply to: SA124B vs BB60D #255078
    AndrewAndrew
    Moderator

    Hi Jom, can you please email us at support@signalhound.com?

    In that email, please provide as much detail in your measurement needs as you can, for instance

    – defining what an HPA is.
    – What are the inputs and outputs of your system?
    – What are you trying to achieve.
    – Anything else you think is relevant to your application.

    This will allow us to provide the best recommendations.

    We look forward to your respnse.

    • This reply was modified 5 months, 3 weeks ago by AndrewAndrew.