The brand-new PN400 Phase Noise Test System is available for purchase! →

Volker

Forum Replies Created

Viewing 20 posts - 1 through 20 (of 22 total)
  • Author
    Posts
  • in reply to: Spike’s phase noise analysis is now a paid add-on! #246683
    Volker
    Participant

    Dear Signalhound,

    looks like the standard phase noise measurement is gone…

    Can you please confirm that by start using version 4.0 this feature is not usable anymore without license.
    In this case you should mark this feature as payable option in the feature list of Spike

    in reply to: indicator for Reference Level Offset #242515
    Volker
    Participant

    Thanks.

    I think it would sometimes also be very interesting to have an information what attenuator is switched in when the unit is set to auto atten. Especially in noise figure measurement.

    Volker

    in reply to: Noise figure measurement #242274
    Volker
    Participant

    HP 346 B is a good but still expensive noise source.
    Works well with SM200B. As most noise sources this one needs a 28V DC powersupply. You have to attach an external power supply.

    the best ?

    SM200B and a ( any ) noise source is a very good way to get an overview for noise figure measurements. Be carefull with cheap noise sources from the internet. Most of them have plenty of power and may damage your analyser.

    I have also used one of the 20 EUR noise sources from the internet. All power stages have to be removed. Then it can be calibrated against a known noise source ( HP346B and Eaton 2075 noise gain analyser ). Works well up to 2GHz with comparable results to HP346B. Still frequency range of HP346B is far beyond the home brew one.

    Keep in mind : noise figure measurements are not as trivial as it looks.

    in reply to: SCPI reading the avg power value in zero span #242082
    Volker
    Participant

    Thanks,

    this is working fine.

    in reply to: SCPI reading the avg power value in zero span #242065
    Volker
    Participant

    Hi Andrew,

    in the release notes from 3.8.0 you wrote:
    > SCPI command FETCH:ZS? can now return average power.

    I have tried but could not get the expected result.

    Parameter 1 delivers IQ Values that can be used with this formular: 10*log10(mean(I^2 + Q^2))

    Parameter 2 delivers the number of IQ values

    looks like no other paramters are implemented.

    Can you please explain how FETCH:ZS? retrieves the avg power value in zero span mode.

    Volker

    in reply to: SCPI reading the avg power value in zero span #241723
    Volker
    Participant

    Hi Andrew,

    thanks for your explanation of the calculation the amplitude value from the I/Q data. Still it would be nice to fetch the value from the analyser. The workaround generates lots of data, especially if you need many amplitude values…

    Yes, I would be interested to have a SCPI command to retrieve it. It makes the resulting script much more easy.

    in reply to: Noise Figure Feature Request #238693
    Volker
    Participant

    Hi Andrew,

    of course a standart noise source needs 28V. There is no need to drive it direct. It would be rather easy to use a trigger signal to drive some switching logic. Just need a signal and maybe a delay to start the next measurement.

    Volker

    in reply to: Noise Figure measurement with Spike #238411
    Volker
    Participant

    It would be a nice feature if you would take the attenuator (if mounted) into account.
    Since this is not really important for NF it would give the right values in the gain plot for documentation purpose. I was thinking about using the “reference level offset” but I think it is not the best solution. Maybe a field “loss after dut” will be better

    in reply to: noise figure measurement #238171
    Volker
    Participant

    Thanks for the improvement in 3.5.13.
    Good Job.

    in reply to: noise figure measurement #238000
    Volker
    Participant

    Shame on me, I had a wrong manual.

    Compliments to you for intuitive implementing of this feature. I will check again with “4.12.5 Limitations” in mind

    in reply to: noise figure measurement #237994
    Volker
    Participant

    Hi Andrew,

    I spent a little more time then I wanted. The function is working nicely and is producing results that are close to what the Eaton is measuring. I have seen peeks in NF at about 850MHz, but this is only with one DUT and I have no idea why…

    Exporting the complete hot/cold data could probably be also a nice feature.

    An application note of how the measurement is done could also be very helpfull. Especially what is done with the input attenuator and the preamp. That also leads to the question what NF the BB60C has. I have not found any informations about this.

    One more automation step could be evaluating the optimal RefLevel, but the first one is definately temporary storing the noise source
    cal to skip steps 1 and 2 of the four steps when you want to repeat the measurement

    Thanks for this feature.

    Volker

    in reply to: noise figure measurement #237951
    Volker
    Participant

    Hi Andrew,

    thanks for the information about the SA44/124.

    I played around with the new NF function and compared this with the values I get from an Eaton 2075. I have a HP 346B and a selfmade noise source that are reproducing realistic values. In the frequency range of about 800 to 1000 MHz I am always getting values to high. The rest is really close to what is expected. I will invest some more time next week and have a closer look.

    For my last question:
    There was a interesting article here

    https://www.mwrf.com/technologies/test-measurement-analyzers/article/21845669/measure-noise-without-a-calibrated-source

    I have not tried this, but the steps to use this measurement method are very close to yours.

    Best,

    Volker

    in reply to: Limit Lines in 3.5.10 #237866
    Volker
    Participant

    Hi Andrew,

    nothing to apologize for. Thanks for this great product and your great support.

    Volker

    in reply to: Limit Lines in 3.5.10 #237853
    Volker
    Participant

    Hi Andrew,

    just found out that the position of the limit lines depend on the RBW. By increasing the RBW the limit lines are shifting left.

    Volker

    in reply to: Limit Lines in 3.5.10 #237851
    Volker
    Participant

    Hi Andrew,

    just found the time to come back to the problem. It is much better in 3.5.11 but still not correct. Limitlines are slightly to low in frequency, but this is working.

    Limitlines should be right on grid.

    Best,

    Volker

    Attachments:
    You must be logged in to view attached files.
    in reply to: Limit Lines in 3.5.10 #237498
    Volker
    Participant

    Hi Andrew,

    thanks for the response. I still have an older version of spike and I can wait for the fixed one.

    Best,

    Volker

    in reply to: IP2 & IP3 CALCULATION IN SPIKE #236822
    Volker
    Participant

    I think there is no need to automatically calc IP3 or IP2. But for documentation it would be very nice to have a display field with a title where you can do marker math and place some comments

    in reply to: Tracking Generator #236816
    Volker
    Participant

    Justin,
    that is good to hear. Sweeptime is normally not what you need while using a TG with 201 points. Adapted frequency rage would be nice. Why not using the VSGs as TG. That would be a nice bundle even if the actual VSG is not covering the full frequency range.

    in reply to: Limit Lines in Spike 3.1.12 #158141
    Volker
    Participant

    Hi Andrew,

    limit testing in version 3.1.14 is now nearly working as expected. Thank you for great service.

    Still there is something strange. In the picture you will see the two lower limit lines ( red needles ) that will check for minimum power level of two carriers. They should come up to +22 dBm. Instead they just show somthing about +18 dBm.

    It is working when you fall below the graphical line, but the line is not corresponding the values in the Limit Line Table.

    Limit2.jpg should not show pass !

    Volker

    Attachments:
    You must be logged in to view attached files.
    in reply to: Limit Lines in Spike 3.1.12 #157922
    Volker
    Participant

    Hi Andrew,

    I think you are right about the RBW. If I come down to a RBW of about 300 Hz it is working. Since this was for testing and quick checking of IP3 of an amplifier, I increased RBW to 5kHz. In this case I can see the two main carriers. With the low RBW they hide behind the limitlines ( picture lowrbw.jpg ).

    If I increase the RBW to 3kHz it will pass ( picture rbw3khz.jpg ).
    It will fail at a summ level somewhere below -73 dBm.

    It gets worse if you increase the span…

    Anyhow this was working in Version 3.1.11.
    So the algorithm of checking has changed. If it is like your wrote that you are checking single samples against the limit you may run into trouble when you have a good generator. Should you not check against a interpolated line between two samples ?

    I really hope you can get it running, because I ever wanted the limit lines stored with the preset as it is first done in this version.

    The limits I used:
    249.8,-200
    249.8995,-200
    249.9000,-60
    249.9005,-200
    250.0995,-200
    250.1000,-60
    250.1005,-200
    250.2000,-200

    Best

    Volker

    Attachments: